Post by wewill on Mar 9, 2012 16:47:41 GMT -5
My 4170 ran great yesterday. However, today, when I tried to measure a vertical antenna, the SWR stayed constant across the top of the screen. (the Theta curve looked fine.)
I tried different custom calibrations, rebooting the computer and cycling the AIM off and on. But nothing helped.
Next I did another custom calibration on 18 inches of new RG-213 which ran with no error messages. But, when I put a load of 100 Ohms on the end of the coax (expecting an SWR of 2:1) I again got a constant and infinite SWR.
What am I doing wrong?
Thanks for your help.
|
|
|
Post by Bob on Mar 9, 2012 18:36:20 GMT -5
Your procedure sounds reasonable.
Make sure Zref is set to 50 ohms. Click Setup->Enter Zo Ref if it needs to be changed.
To make sure the hardware is still working, do a standard cal with the loads connected directly to the RF connector. After cal is complete do a scan from 1 to 20 MHz with a step size of 0.2MHz. This should give a Zmag of 50 ohms and SWR = 1.0. If the cal resistor is not 50 ohms, that's ok. The SWR will read Rcal/50.
-- 73/Bob
|
|
|
Post by wewill on Mar 10, 2012 16:17:51 GMT -5
Hi Bob,
The Zref was the problem. When I went to Setup>Enter Z0 Ref, it read something like 50 + j338136628000000000000000000000. When I set it to 50, everthing worked fine.
Thanks again for the quick and accurate fix.
|
|