Post by ddhlv on Nov 25, 2010 11:15:54 GMT -5
Happy Holiday to all:
I am have a problem with reading my SWR, Refl Coef, % refl power, and Return Loss. Each scan the number stay the same. I am running WinXP SP3 on each computer, one is a laptop, the other is a desktop. Example below.
SWR= 1.#QO Refl Coef= 1.#QO %refl power= 1.$ Return Loss- -1.#.Jdb
Equivalent Circuit numberbers are readable. Thanks Dennis
|
|
|
Post by Bob on Nov 25, 2010 12:35:59 GMT -5
Hello Dennis,
Does the scan of an antenna or a component look reasonable? After calibrating, scan the load resistor that was used for calibration. It should produce a nice flat line. If it's a 50 ohm resistor, the SWR will be 1.0 and the return loss will be greater than 30 dB.
Make sure the plots for SWR and return loss are enabled.
-- 73/Bob - W5BIG
|
|
ddhlv
New Member
Posts: 5
|
Post by ddhlv on Nov 25, 2010 15:42:29 GMT -5
I understand Bob. I did not have my number enter correctly for Zo. All numbers are readable. Thanks
dennis lee valentine, k8dlv
|
|