Post by charlie on Oct 25, 2016 10:00:32 GMT -5
Hi! I just got my AIM 4300 and am in a Steep learning curve. From what I have seen so far, this is an excellent analyzer.
I am running the software (AIM 910B) on a Win 10 Dell. I encountered no problems hooking it up and getting it to talk.
However, on my first try to calibrate, I got a "Invalid Calibration Temperature Error" Error 26803. Did a little looking and found a value close to what I saw (~7300), I reset this value to 73 degrees and it took off. Is this OL or do I need to do something else?
Second, and this is probably my infamiliarity with the software. Where is the Data Log Button?
Thanks Guys! Charlie
|
|
|
Post by Bob on Oct 25, 2016 12:34:11 GMT -5
Hi Charlie,
Can you please send me a screen shot showing this error message? This might indicate a problem in the cal data file, like a file with an obsolete format. After doing a new calibration, it should be ok.
-73/Bob
bob@w5big.com
|
|
|
Post by charlie on Oct 25, 2016 13:50:27 GMT -5
Bob,
Thanks very much! It looks like if fixed my problem, I reran a new calibration and got no problems. I will send a copy of the calibration file (recalibration). It gives me a flat response on the 50 ohm load, so all good
I think I figured the data save out. I save the graph and then import as a CSV. Looks great!
73 Charlie
|
|